嗨,在N6705B机箱中,是否可以将N6784A模块远程编程为自动测距状态?
此功能有一个手动选择(可通过前面板[属性]按钮访问),完全按预期工作。 但是,编译帮助非常清楚地说明命令SENSe:VOLTage [:DC]:RANGe:AUTO仅用于N6781A和N6782A - 实际上它不适用于N6784A。 那么,我是否遗漏了某些内容,或者此功能是否未实现远程控制? 感谢任何帮助或澄清,-Mark Royer 以上来自于谷歌翻译 以下为原文 Hi, Is it possible to remote program the N6784A module for auto ranging state while in an N6705B chassis? There is a manual selection for this feature (accessible via the front-panel [Properties] button) that works exactly as expected. However, compiled help very clearly states that the command SENSe:VOLTage[:DC]:RANGe:AUTO is for the N6781A and N6782A only - and indeed it doesn't work on the N6784A. So, am I missing something or is this feature just not implemented for remote control? Thanks for any help or clarifications, -Mark Royer |
|
相关推荐
2个回答
|
|
嗨Mark,远程自动量程确实只适用于N6781A和N6782A。
前面板自动量程很慢,基本上都在进行测量,如果超出范围,它会提升范围。 虽然它在前面板上并不明显,因为它具有更新速率的好处。 通过界面,您可以在每次测量时以5 us的速度进行测量样本,因此我们无法在那里进行测量。 对于那个很抱歉。 以上来自于谷歌翻译 以下为原文 Hi Mark, The remote auto range is indeed only available on the N6781A and N6782A. The front panel autorange is slow and is basically doing measurements and if it over ranges, it steps up the range. It is not noticeable on the front panel though because it has the benefit of the update rate. Over the interface, you can do measurement samples as quickly as 5 us per measurement so we cannot do that there. Sorry about that. |
|
|
|
张友芬1 发表于 2018-9-20 16:42 谢谢,马特,回复。 我明白你在说什么。 然而,我们正在努力使用通用(本土)软件应用程序进行非常低电流(大约10 uA)的自动测量,该应用程序不会提前知道负载。 因此,它无法猜测所需的范围,当然,在更高的范围内,精度对于非常低的电流是不可接受的。 有关如何绕过这个的任何想法? 我的最后一个解决方案是让测试技术手动设置范围,这取决于预期的负载 - 但是,这确实是一个不太令人满意的解决方案。 再次,感谢Matt,-Mark R.嗯...没关系...... :)重新阅读你的答案,它发生在我身上我应该可以做与前面板自动调整相同的事情 - 如果软件检测到 超范围然后加强范围再试一次。编辑:mroyer36于2015年9月24日下午3:56 以上来自于谷歌翻译 以下为原文 Thanks, Matt, for the reply. I understand what you're saying. However, we are struggling making automated measurements of very low current (circa 10 uA) with a general purpose (home grown) software app that doesn't know the load ahead of time. So, it can't guess the range required and on the higher ranges the accuracy is unacceptable for very low currents, of course. Any thoughts on how to get around this? My last ditch solution will be to have the test tech set the range by hand, depending on the anticipated load - but, that's really a less satisfactory solution. Again, thanks Matt, -Mark R. Ummm... nevermind... :) Re-reading your answer, it occurs to me I should be able to do the same thing as the front-panel autoranging - if the software detects an overrange then step up the range and try again. Edited by: mroyer36 on Sep 24, 2015 3:56 PM |
|
|
|
只有小组成员才能发言,加入小组>>
1287 浏览 0 评论
2375 浏览 1 评论
2194 浏览 1 评论
2064 浏览 5 评论
2951 浏览 3 评论
1111浏览 1评论
关于Keysight x1149 Boundary Scan Analyzer
753浏览 0评论
N5230C用“CALC:MARK:BWID?”获取Bwid,Cent,Q,Loss失败,请问大佬们怎么解决呀
926浏览 0评论
1287浏览 0评论
小黑屋| 手机版| Archiver| 德赢Vwin官网 ( 湘ICP备2023018690号 )
GMT+8, 2024-12-23 23:00 , Processed in 1.575432 second(s), Total 80, Slave 64 queries .
Powered by 德赢Vwin官网 网
© 2015 bbs.elecfans.com
关注我们的微信
下载发烧友APP
德赢Vwin官网 观察
版权所有 © 湖南华秋数字科技有限公司
德赢Vwin官网 (电路图) 湘公网安备 43011202000918 号 电信与信息服务业务经营许可证:合字B2-20210191 工商网监 湘ICP备2023018690号