* PNA型号* E8363C;
A.09.42.22;
选项010,014,P02 *目标:*表征100W放大器的增益,绝对输出功率,VSWR,增益压缩。
*测试设置:*类似于应用笔记1408-10的图4,除了我不需要前置放大器和我的输出测试耦合器是30 dB。
我也没有在rcvr B in和Cplr
ARM之间使用10 dB。
所需的端口1功率(DUT输入)将从-25dBm到0dBm。
* PNA配置:* 101分,IFBW = 1kHz,Avg = 7 *问题1:*校准后,S21代替DUT不平坦(> 1.5 dB)我首先使用功率计/传感器在端口上进行功率校准
1(-25dBm至0 dBm,步长为1 dB)。
然后我用eCal模块(N4691-60004)更换我的DUT并进行全2端口S-param校准(端口功率为-10 dBm)。
接下来我把_thru放入我的DUT_然后看S21(dB)并且没有得到我期望的平坦度。
我尽量减少测试电缆的移动,并小心所有连接。
如果我做了响应直通,那么我可以获得平坦度我有不同的IFBW,平均值,初始端口功率水平,测试电缆,测试耦合器,具有类似的结果。
*问题2:*我可以显示DUT的绝对功率(抵消测试耦合器和电缆的输出损耗)吗?
为了查看DUT的绝对输出功率,我正在测量接收器B / 1.0(源= P1)。
我可以偏移此数据(通过频率)来显示我的DUT的
电源输出吗?
*问题3:*您能否确认我的PNA配置不允许内部调整接收器B的衰减?
另外,我是否需要在外部添加衰减(cplr arm到RCVR B in)以减少Receiver B压缩?
任何帮助表示赞赏。
以上来自于谷歌翻译
以下为原文
*PNA Model* E8363C; A.09.42.22; Op
tions 010, 014, P02
*Goal:* characterize Gain, Abs Output Power, VSWR, Gain Compression of a 100W amplifier.
*Test setup:* similar to Fig 4 of App Note 1408-10 except I do not need the preamp and my output test coupler is 30 dB. I also am not using the 10 dB between the rcvr B in and Cplr Arm.
Desired Port 1 power (DUT input) will be from -25dBm to 0 dBm.
*PNA config:* 101 pts, IFBW = 1kHz, Avg = 7
*Issue1:* S21 with thru in place of DUT is not flat (>1.5 dB) after calibration
I first conduct a power calibration with my power meter/sensor on Port 1 (-25dBm to 0 dBm in 1 dB steps).
Then I replace my DUT with the eCal module (N4691-60004) and do a full 2 port S-param calibration (at port power of -10 dBm).
Next I put a _thru in for my DUT_ and look at S21 (dB) and do not get the flatness I would expect. I try to minimize test cable movement and am careful with all connections.
If I do a Response Thru cal then I can get flatness < 0.05dB, but thought full 2 port would be the best cal and also want to meas other S-params.
I have varied IFBW, Averaging, Initial port power level, test cables, test coupler, with similar results.
*Issue2:* Can I display the absolute power at the DUT (offset the output losses of the test coupler and cables)?
In order to see the absolute output power of the DUT, I'm measuring Receiver B/1.0 (source = P1). Can I offset this data (over freq) to display the power out at my DUT instead?
*Issue 3:* Can you confirm that my PNA configuration does not allow for internal adjustment of the Receiver B attenuation? Also, will I need to add attenuation externally (cplr arm to RCVR B in) in order to reduce the Receiver B compression?
Any help is appreciated.