引入了NIB30A网络分析仪,GPIB设置已更改。
没有内部gpib接口配置来从内部窗口控制仪器。 外部GPIB设置为7xx地址,副本为14xx。 io配置有一个GPIB0和GPIB1而不是HPIB7,gpib0设置为地址7,GPIB1设置为33.有没有办法解决这个问题而不进行恢复? -大卫 以上来自于谷歌翻译 以下为原文 Have brought in a N5230A network analyzer that the GPIB Settings have been changed. There is no internal gpib interface configure to control the instrument from the internal windows. The external GPIB is set to 7xx address with a duplicate at 14xx. The io config has a GPIB0 and GPIB1 instead of HPIB7 with the gpib0 set to address 7, and GPIB1 set to 33. Is there a way of resolving this issue with out doing a restore? -David |
|
相关推荐
2个回答
|
|
恢复可能是确保所有内容都恢复正常设置的最安全的方法。
根据所有内容的状态,手动重建配置可能会变得更加困难(如果不是不可能)。 我建议恢复。 以上来自于谷歌翻译 以下为原文 The restore is probably the safest way to make sure that everything comes back setup properly. Depending on the state of everything, it could get harder (if not impossible) to rebuild the configuration manually. I would recommend the restore. |
|
|
|
好吧,我得到了解决方案,星期一,当论坛失败时,考虑到它是一个借用机器,对恢复犹豫不决。 但是在IO库中,config能够双击左侧的内部接口,并创建了一个HPIB7引用。 删除了设置为地址33的GPIB1并将gpib0移动到地址14。 以上来自于谷歌翻译 以下为原文 Well I got it fixed, Monday while forums were down, was hesitant on the restore considering it a loaner machine. But in IO library config was able to double click the internal interface on the left side and it created a HPIB7 reference. Removed the GPIB1 that was set to address 33 and move the gpib0 to address 14. |
|
|
|
只有小组成员才能发言,加入小组>>
1279 浏览 0 评论
2371 浏览 1 评论
2187 浏览 1 评论
2061 浏览 5 评论
2945 浏览 3 评论
1099浏览 1评论
关于Keysight x1149 Boundary Scan Analyzer
748浏览 0评论
N5230C用“CALC:MARK:BWID?”获取Bwid,Cent,Q,Loss失败,请问大佬们怎么解决呀
917浏览 0评论
1279浏览 0评论
小黑屋| 手机版| Archiver| 德赢Vwin官网 ( 湘ICP备2023018690号 )
GMT+8, 2024-12-21 18:14 , Processed in 1.585086 second(s), Total 78, Slave 62 queries .
Powered by 德赢Vwin官网 网
© 2015 bbs.elecfans.com
关注我们的微信
下载发烧友APP
德赢Vwin官网 观察
版权所有 © 湖南华秋数字科技有限公司
德赢Vwin官网 (电路图) 湘公网安备 43011202000918 号 电信与信息服务业务经营许可证:合字B2-20210191 工商网监 湘ICP备2023018690号